KBA Technotrans Problem

Discussion in 'KBA Printing Presses' started by John Samuel, Feb 20, 2020.

  1. John Samuel

    John Samuel New Member

    Joined:
    Feb 2020
    Messages:
    2
    Location:
    Philippines
    Good day to all mamabers,

    I have a problem with our 3 technotrans unit with the alcomart AZR .

    The units already prompting an error of IPA>18% we already done what is indicated on the manual

    but the error still occurs. We also try not pump it with alcohol only water and solution but its still

    indicating IPA>18%. Hope any one has experienced this before.

    Thanks and good day.
     
  2. CHOCK

    CHOCK Senior Member

    Joined:
    Dec 2016
    Messages:
    3,005
    Location:
    Managua,Nicaragua
    Hi sir
    Attached is there the sheet of calibration sensor firt in the intruction m

    Let me know

    Blessing
     

    Attached Files:

  3. John Samuel

    John Samuel New Member

    Joined:
    Feb 2020
    Messages:
    2
    Location:
    Philippines
    Is this to be done on the alcosmart unit or on the console (technotrans remote).
    Thanks also for the reply
     
  4. msaddek

    msaddek Member

    Joined:
    Nov 2019
    Messages:
    9
    Location:
    Morroco
    Is this to be done on the alcosmart unit.
     
  5. Gregory Ross

    Gregory Ross Member

    Joined:
    Sep 2020
    Messages:
    11
    Location:
    Fl
    Don,t know what to say, haven't run alcohol in 25 years
     
  6. CHOCK

    CHOCK Senior Member

    Joined:
    Dec 2016
    Messages:
    3,005
    Location:
    Managua,Nicaragua
    Hi sir:

    Please go to the start menu:

    Go to level 2 for diagnosis troubleshooting.

    When the star key is pressed the calibration values(gain, offset) will be displayed
     
  7. Adil

    Adil Senior Member

    Joined:
    Nov 2021
    Messages:
    1,088
    Location:
    Casablanca
    Hello, if you have not yet solved the problem, try changing the membrane, the filter and the seal on the alcosmart (if the filter is saturated with alcohol it always gives you a bad reading). It is a flat filter of one color white.

    If it does not work with this basic maintenance then you have to change the gas detector inside and calibrate it afterwards; this intervention is done at the technotrans in your country.

    The duration of this operation does not exceed 15 days.
     
    Last edited: Sep 7, 2022
  8. georgbob

    georgbob Member

    Joined:
    Jul 2019
    Messages:
    26
    Location:
    Hungary
    clean the protective plate, wash the filter with water
     
  9. Moni Patel

    Moni Patel New Member

    Joined:
    Dec 2021
    Messages:
    4
    Location:
    India
    Did you find a solution to your problem? Even I am having the same problem. Kindly guide me on the process to solve this problem.
     

  10. Adil

    Adil Senior Member

    Joined:
    Nov 2021
    Messages:
    1,088
    Location:
    Casablanca
    Hi sir
    Please do that .
    Hello, if you have not yet solved the problem, try changing the membrane, the filter and the seal on the alcosmart (if the filter is saturated with alcohol it always gives you a bad reading). It is a flat filter of one color white.

    If it does not work with this basic maintenance then you have to change the gas detector inside and calibrate it afterwards; this intervention is done at the technotrans in your country.

    The duration of this operation does not exceed 15 days.
    Good luck
     