KBA Technotrans Problem

John Samuel

New Member
Joined
2020
Posts
2
Geo
Philippines
Good day to all mamabers,

I have a problem with our 3 technotrans unit with the alcomart AZR .

The units already prompting an error of IPA>18% we already done what is indicated on the manual

but the error still occurs. We also try not pump it with alcohol only water and solution but its still

indicating IPA>18%. Hope any one has experienced this before.

Thanks and good day.
 
Hi sir
Attached is there the sheet of calibration sensor firt in the intruction m

Let me know

Blessing
 

Attachments

  • technotran calibration.png
    technotran calibration.png
    187.4 KB · Views: 1,347
  • Hi sir
    Attached is there the sheet of calibration sensor firt in the intruction m

    Let me know

    Blessing
    Is this to be done on the alcosmart unit or on the console (technotrans remote).
    Thanks also for the reply
     
    Good day to all mamabers,

    I have a problem with our 3 technotrans unit with the alcomart AZR .

    The units already prompting an error of IPA>18% we already done what is indicated on the manual

    but the error still occurs. We also try not pump it with alcohol only water and solution but its still

    indicating IPA>18%. Hope any one has experienced this before.

    Thanks and good day.
    Don,t know what to say, haven't run alcohol in 25 years
     
    Hi sir:

    Please go to the start menu:

    Go to level 2 for diagnosis troubleshooting.

    When the star key is pressed the calibration values(gain, offset) will be displayed
     
    Hello, if you have not yet solved the problem, try changing the membrane, the filter and the seal on the alcosmart (if the filter is saturated with alcohol it always gives you a bad reading). It is a flat filter of one color white.

    If it does not work with this basic maintenance then you have to change the gas detector inside and calibrate it afterwards; this intervention is done at the technotrans in your country.

    The duration of this operation does not exceed 15 days.
     
    Last edited:
  • Did you find a solution to your problem? Even I am having the same problem. Kindly guide me on the process to solve this problem.
     
    Hi sir
    Please do that .
    Hello, if you have not yet solved the problem, try changing the membrane, the filter and the seal on the alcosmart (if the filter is saturated with alcohol it always gives you a bad reading). It is a flat filter of one color white.

    If it does not work with this basic maintenance then you have to change the gas detector inside and calibrate it afterwards; this intervention is done at the technotrans in your country.

    The duration of this operation does not exceed 15 days.
    Good luck
     
    Forum posts represent the experience, opinion, and view of individual users. The Color Printing Forum does not necessarily endorse nor share the view of each individual post. When making any potentially dangerous or financial decision, always employ and consult appropriate professionals. Your circumstances or experience may be different.