No sir, you need test all the Power supply for Vdc. NOT only +5vDC 2) Measure: a) -8vdc b)+8Vdc c) +12Vdc ) 3)How you did test the I tested...
ok look for the following: 1) test for dry +col solder joint 2) you did NOT reply about my question : However if the NTK ARE IN Green light, NOT...
Hi sir: 1)Test all Ribbon cables or flat cables, for : a) bad contacts b) loose cables c) dirty d) oxide e) broken) 2) Edge connectors are a...
Hi sir: Please review: Please check the function and value of the potentiometer, look for his schematic diagram electrical
Hi sir: 1) You must measure the power supply for the CPU : Vcc ! other wise the Reset will NOT effective.(attached is there example for CPU 8085...
ok sir Let me know. Blessing
Hi sir: 1)Measure Clock for CPU: Check the frecuency with frequency counter or a DMM included frecuency counter( like Fluke DMM) 2)Reset CPU:...
Hi sir: I strong recommend you, that you measure the input of machine in the section power pack : 220Vac or 380Vac for Feeder Transformer, they...
Hi sir: Please do Reset in the Small button in the electronic board CP2 or CP4 . The button reset, has small LED V1, that must Turn ON, when is...
Thank you sir by his Words. Am here to try to Help, anybody.
Please take t steps indicate before, Did you do Sir? Otherwise you would take around in the same problem.
Replace Display please. Blessing
The dry+ colder joint Solder. Blessing.
Hi sir: Please make this measure: 1)Key-board and display BT: + 9 Volt D.C. Tolerance: 8,4 - 9,6 Volt D.C. 2) Key board BR: + 5 Volt D.C....
Please take t steps indicate before Otherwise you would take around in the same problem.
Hi sir; You must go to the shop machine, the motor, for maintenance time, and exchange the Encoder Every electric motor, must have maintenance...
Hi sir Am happy that you fix it , the problem:) As i did indicate,;2) Normally you can touch RESET the error Motor button
Hi sir: Let me know, what are you doing the press? is running , or inching, , reverse possible, forward possible, alarm continues to sound. 2)...
Hi sir; You must test very well the Pumps, the facts the c. breaker is tripped is indicative the pumps(mechanical, specially winding re'heating...
Hi sir; Please do my steps very carefully, otherwise, you will take around the same problem. Blessing